(1) Background: Continuous monitoring of tacrolimus (TAC), mycophenolic acid (MPA), and creatinine (Cre) after renal transplantation is vitally important. In this study, we developed a new method based on volumetric absorptive microsampling (VAMS) combined with Ultra Performance Liquid Chromatography−Tandem Mass Spectrometry (UPLC-MS/MS) to simultaneously quantify three analytes including TAC, MPA, and Cre in whole blood. (2) Methods: The VAMS-based UPLC-MS/MS assay used a shared extraction and a single injection to simultaneously quantify the included TAC, MPA, and Cre. Development and validation were carried out following the Food and Drug Administration and European Medicines Agency guidelines for the validation of bioanalytical methods. Moreover, clinical validation for the three analytes was performed in both dried blood spot (DBS) and VAMS. Furthermore, a willingness survey was conducted using the system usability scale (SUS) for renal transplant recipients. (3) Results: The assay was successfully validated for all analytes. No interference, carryover, or matrix effects were observed, and extraction recoveries and process efficiencies were >90.00%. Analysis was unaffected by hematocrit (0.20~0.60, L/L) and anticoagulants (EDTA-2K). Dried VAMS samples were stable for 7 days at ambient temperature and stable for at least 1 month at −20 °C. During clinical validation, the measured TAC, corrected MPA, and Cre concentrations of VAMS samples met the analytical standards (95.00%, 88.57%, and 92.50%). When more stringent clinical acceptance criteria were set, the results obtained by VAMS (90.00%, 71.43%, and 85.00%) better than DBS (77.50%, 62.86%, and 70.00%). Compared with DBS, the survey found that renal transplant recipients are more inclined to use VAMS. (4) Conclusions: A robust extraction and UPLC-MS/MS analysis method in VAMS tips was developed and fully validated for the simultaneous quantification of TAC, MPA, and Cre concentrations. This method provides analytical support for the one-sample remote monitoring of both immunosuppressive drug concentrations and renal function in allo-renal recipients. Based on the good consistency between this method and the routine detection of venous blood samples and higher patient satisfaction than DBS, we believe that VAMS sampling can be a better alternative to venous whole-blood sampling.
Read full abstract