PurposeIn Uganda, two-dimensional (2D) radiotherapy treatments have been in use since the establishment of radiotherapy in 1995. Preliminary investigations of treatment records in November 2019 showed evidence of gaps requiring urgent attention. The purpose of this study was to improve the safety of the treatments. MethodsRecords of 1164 patients treated in 1387 courses (1412 sites) on Cobalt-60 units were reviewed todetermine the frequency and dosimetric implications of events that occurred at different stepsof the radiotherapy process. The results were presented and discussed with the differentprofessionals for learning purposes. ResultsMost common dosimetric eventswere omission of block tray, bolus and couch transmission factors in time calculations, incorrect field sizes and depths, wrong beam weighting, independent calculations and prescription doses contributing 28.6 %, 10.1 %, 6.0 %,11.9 %, 10.1 %, 5.4 %, 4.8 % and 8.9 % to the 168 observed errors. Comparison of the calculated treatment doses with the prescribed doses showed that 88 % of the 1412 sites were treated with radiation doses within an accuracy of ± 5 %. However, an analysis of the evolution along the years demonstrated an improvement from 82.8 % in 2018 to 86.1 % in 2019, and 93.2 % in 2020. Most common procedural events were incomplete setup instructions and missing patient data in the record and verify system of the Co-60 units for 57 % and 60.1 % of the 1164 patients. ConclusionsOpportunities for improvement of safety in the delivery of radiotherapy treatments were identified. Learning from these past errors should raise awareness in the team leading to a safer treatments.