Total body irradiation before bone marrow transplantation for hematological malignancies using Radixact, a high-precision radiotherapy machine, can potentially reduce side effects and the risk of secondary malignancies. However, stable control of couch speed is critical, and direct assessment methods outlined in quality assurance guidelines are lacking. This study aims to develop a real-time couch speed verification system for the Radixact. The developed system used a linear encoder to measure couch speed directly. Accuracy was verified via a linear stage, comparing measurements with a laser distance sensor. After placing a phantom simulating the human body on the Radixact couch, the couch speed was verified using predefined speed plans. Operating the linear stage at 0.1, 0.5, and 1.0mm/s revealed that the maximum position error of the developed verification system compared to the laser distance sensor was nearly equivalent to the distance resolution of the system (0.05mm/pulse), with negligible average speed error. When the Radixact couch operated at 0.1, 0.5, and 1.0mm/s, the values obtained by the verification system agreed with the theoretical values within the sampling period (0.01s) and distance resolution (0.05mm). The verification system developed provides real-time monitoring of the speed of the Radixact table, ensuring treatment effectiveness and patient safety. It would guarantee the couch speed's soundness and contribute to the "visualization" of safety.