Two procedures have been proposed and compared for separation and preconcentration of trace amounts of manganese, copper, and zinc in saline samples. The procedures are based in the use of Amberlite XAD-7 and an anion-exchanger Dowex 1X8-50 loaded with Alizarin Red S (ARS). In order to obtain quantitative recoveries of metal ions, various experimental parameters such as pH, shaking time, sample volume, amounts of solid phase, effects of concomitants, capacity and cations desorptions from solid phases were optimized. For Amberlite XAD-7 impregnated with ARS (XAD 7-ARS), Mn, Cu, and Zn were quantitatively retained, in the pH range 8.5-10.0, by using 0.50 g of solid phase, stirring time of 5 min and a total mass up to 200 µg of each cation. The sorbed elements were subsequently eluted with 5 mL of 3 mol L-1 HNO3, with recovering over 95%, and a fifty-fold preconcentration factor for Zn and a ten-fold preconcentration factor for Mn and Cu were obtained. For Dowex 1X8-50 impregnated with ARS (Dowex-ARS), Cu and Mn were quantitatively retained in the pH range 8.1 to 9.0, and Zn pH 8.1, by using 0.75 g of solid phase, 90 min of stirring time and a total mass up to 50 µg of each cation. The sorbed elements were subsequently eluted with 20 mL of 2 mol L-1 HNO3, and a five-fold preconcentration factor to Cu and Zn was obtained. The precision of the procedure was determined by running 10 replicate samples, each one containing 250 µg L-1 of each element and the relative standard deviations were 0.2% (Cu), 0.4% (Mn), and 0.4% (Zn), to XAD 7-ARS and 0.3% (Cu), 0.5% (Mn), and 0.3% (Zn), to Dowex-ARS. The procedures were used for determining of Mn, Cu, and Zn in physiological solutions and seawater samples, from Salvador-Bahia. The analyte addition technique was used and the recoveries obtained (³ 95%) revealed that the proposed procedure shows good accuracy and precision.
Read full abstract