Amethod for the determination of zinc, 1ead and cadmium in hot spring waters by atomic absorption spectrometry has been studied The instrument used was Shimadzu Multichannel Atomic Absorption-Flame Spectrometer, Model MAF-1 The sensitivity obtained by means of the stabilization of electric power and the scale expansion of the recoder readings was sohigh that the sample s 1utions were directly at mized int the acetylene-air ameThe condition recommended for the determinatio:n of zinc, 1ead and cadmium were as follows, wavelength;2139 A, 2170 A a:nd 2288 A, ho110w cathode lamp current;6mA 8mA and 6 mA slit width; 0.05 mm, 0.10 mm and O.05 mm, acetylene pressure;0 2kg/ cm2 0 2kg/ cm2 and O 2kg/ cm2, air pressure;1.4 kg/ cm2, 1.4 kg/ cm2 and 1 4 kg/ cm2 beam p slti n;17 mm 16 mm and 20mm, respectively.The presence of aids and coexisting elements did not iRterfere, except a large amount of silicon for zin. This interference could be eliminated through an addition of calcium or lanthanum to the sample solution.The detection limits, determined by means of respective calibration curves, were Zn O.005 ppm, Pb O.Ol ppm and Cd O.OOI C. The method was applied to the determination of the three elements in various types of hot spring waters and the results were satisfactorily coincident with those obtained by other methods.