In patients with refractory atrial fibrillation (AF), atrioventricular nodal (AVN) ablation and permanent pacemaker implantation is recommended. The Micra Transcatheter Pacing System (Micra) is a single chamber leadless pacemaker (LPM) and thus offers the possibility of AV node (AVN) ablation in the same procedure. Pacing threshold (PT) elevation after radiofrequency (RF) ablation is a potential complication. We conducted a single center retrospective cohort study. Patients implanted with a Micra (n = 84) and concomitant or delayed AVN ablation (n = 12) from 2014 to 2022 were included. Two cases of acute Micra PT elevation immediately following RF AVN ablation required device retrieval and implantation of a new Micra. Procedural characteristics and electrophysiological parameters were analyzed, and a computer model was performed to determine factors responsible for acute PT elevations. A total of 84 patients were included. Mean age was 74±10 and 48% were women. Twelve patients (14%) underwent AVN ablation. Two patients had acute PT elevation requiring device retrieval despite no direct contact of the ablation catheter with the Micra. Computer modeling shows that significant dissipated power due to electrical field coupling can occur at the tip or ring electrode if the catheter is not kept at a safe distance (≥35mm) from the Micra when a maximum power of 100W is delivered. Concurrent AVN ablation and Micra implantation is safe in most patients. To prevent acute PT elevation, keeping a safe distance of ≥35mm from the tip and ring electrodes of the Micra and using lower power output may prevent this complication.
Read full abstract