Coronary artery calcification (CAC) is an important indicator of coronary disease. Accurate volume quantification of CAC is challenging using computed tomography (CT) due to calcium blooming, which is a consequence of limited spatial resolution. Ex vivo coronary specimens were scanned on an ultra-high-resolution (UHR) clinical photon-counting detector (PCD) CT scanner, and the accuracy of CAC volume estimation was compared with a state-of-the-art conventional energy-integrating detector (EID) CT, a previous-generation investigational PCD-CT, and micro-CT. CAC specimens () were scanned on EID-CT and PCD-CT using matched parameters (120kV, 9.3mGy ). EID-CT images were reconstructed using our institutional routine clinical protocol for CAC quantification. UHR PCD-CT data were reconstructed using a sharper kernel. An image-based denoising algorithm was applied to the PCD-CT images to achieve similar noise levels as EID-CT. Micro-CT images served as the volume reference standard. Calcification images were segmented, and their volume estimates were compared. The CT data were further compared with previous work using an investigational PCD-CT. Compared with micro-CT, CT volume estimates had a mean absolute percent error of for clinical PCD-CT, for EID-CT, and for previous-generation PCD-CT. Clinical PCD-CT absolute percent error was significantly () lower than both EID-CT and previous generation PCD-CT. The mean calcification CT number and contrast-to-noise ratio were both significantly () higher in clinical PCD-CT relative to EID-CT. UHR clinical PCD-CT showed reduced calcium blooming artifacts and further enabled improved accuracy of CAC quantification beyond that of conventional EID-CT and previous generation PCD-CT systems.