Background: Infusion pumps are critical in delivering fluids, including medications and blood products, in controlled amounts. However, conventional pumps can cause hemolysis and other issues such as flow variations and infection risks, especially during blood transfusions. To address these limitations, a novel cylinder-type infusion pump, the Anyfusion H-100, was developed, which includes a specialized blood transfusion cartridge set that combines syringe and peristaltic infusion methods. This study evaluates the accuracy and hemolysis rates of the Anyfusion H-100 compared to conventional pumps, aiming to confirm its viability as a safe and effective medical device for blood transfusions. Methods: This study evaluated six different infusion rates (10–180 cc/hr) and conducted 57 transfusion trials, 20 of which used a 1:1 blood–saline dilution. Blood transfusion accuracy was measured using research-grade packed red blood cells, and hemolysis rates were assessed before and after transfusion by chi-square tests and independent sample t-tests. Results: Anyfusion demonstrated an average transfusion error rate of 3.77%, compared to 4.00% for the Terufusion, with no statistically significant difference in hemolysis rates (p = 0.697). Bland–Altman plots confirmed their equivalent performance, with hemolysis rates of 0.566 ± 0.095% for Anyfusion and 0.518 ± 0.126% for Terufusion. Conclusions: Anyfusion provides an accurate and reliable blood transfusion performance comparable to that of Terufusion, with no significant difference in hemolysis rates; its integration of syringe and infusion methods shows a potential for safer and more efficient transfusion practices, especially in pediatric and emergency settings.