We present the first dosimetric measurements of single session, multi-target BgRT deliveries using a clinically realistic motion phantom on a research-only version of the RefleXion X1 system. A custom-made anthropomorphic phantom of a human torso with embedded fillable targets mimicking 18F-FDG-avid lesions was used. From the three embedded spherical targets, Target 1 was 26 mm in diameter coupled with a 3D independent respiratory motion with 22 mm range, whereas Target 2 and 3 were 22 mm in diameter and moved with a 1D 5 mm maximum sinusoidal motion. The 18F-FDG concentration in the background cavity of the phantom was 5 kBq/ml, and the targets were loaded with 10:1, 8:1 and 6:1 contrast relative to the background for Targets 1, 2, 3, respectively. Spherical structures were contoured as GTVs (CTV = GTV) and a 5 mm margin was added to create PTVs. Motion extent of the tumors were captured to create biological tracking zones for each target. Treatment plans were generated using a research version of the Reflexion treatment planning software to deliver 8 Gy/fx to the PTVs. The treatment delivery was repeated 2 times, and each time the phantom was refilled according to the plan. PET image evaluation metrics for each of the three targets were also recorded. Target dosimetry was measured using a combination of radiographic film and ion chamber. The maximum distance between the 97% prescription isodose line from the plan and the film measurements was used to characterize the dosimetric accuracy of the tracked deliveries. CTV and PTV min, max, and mean doses measured on film were also recorded for each target. Treatment plans were successfully created with 100% prescription dose coverage to each target loaded with different FDG ratios. Total treatment times for the single-plan, three-target deliveries were less than 80 minutes. PET evaluation metrics at imaging-only and pre-scan, and planning and film dosimetry to the GTV and PTV for each of the three targets is shown in table below (mean ± standard deviation of both deliveries). The CTV dose coverage was maintained for all targets. The shrinkage distance of the 97% prescription dose isodose line on the film plane for all three targets was less than 3 mm for both tests, and ranged from -0.4 to -2.34 mm. These results demonstrate that high tracking accuracy and dosimetric accuracy can be achieved in single session, multi-target deliveries over a range of target-to-background 18F-FDG concentrations and target motion patterns.
Read full abstract