Abstract

The Taiwan Photon Source (TPS) is a third-generation, 3-GeV storage ring with a design current of 500 mA. To monitor the photon beam stability, blade-type X-ray beam position monitors (XBPMs) are installed to record and monitor the photon beam position. The beam position coordinates involve both the X- and Y-axes, and we must consider coupling effects in the XBPM, which have to be reduced for reliability and optimum precision. This study focuses on the front-end (FE) XBPM calibration via four-blade normalization and suppression matrix compensation. The calibration results show that in the optimal measurement range of ±200μm, the coupling effect between the X and Y position is suppressed to less than a 1-μm drift, which satisfies the high-accuracy beam position requirements for user operation. The reliability and precision of the XBPM system improved after applying the serial calibration process developed in this study, and the detailed derivation of each calibration step will be discussed in this paper.

Full Text
Paper version not known

Talk to us

Join us for a 30 min session where you can share your feedback and ask us any queries you have

Schedule a call

Disclaimer: All third-party content on this website/platform is and will remain the property of their respective owners and is provided on "as is" basis without any warranties, express or implied. Use of third-party content does not indicate any affiliation, sponsorship with or endorsement by them. Any references to third-party content is to identify the corresponding services and shall be considered fair use under The CopyrightLaw.