Volatilization of arsenic, selenium and antimony for sample introduction in atomic absorption spectrometry has been performed by pumping an acidic sample through an anion exchanger in the tetrahydroborate (III) form packed as a bed in the liquid channel of a gas-liquid separation membrane cell. The hydrides generated in the heterogeneous reaction between bound tetrahydroborate (III) ions and the analytes are rapidly transferred with the aid of the concomitantly generated hydrogen gas through the gas-permeable membrane into the gas phase and swept to the spectrometer by an additional hydrogen gas flow. This instant transfer of the hydrides to the gas phase kinetically discriminates the reaction of the hydride with metal borides and metal colloids, whose formation by reaction with tetra-hydroborate (III) is slower than the hydride reaction. The susceptibility to interference by transition metal ions is thus markedly reduced, as compared with both batch hydride generation methods and a previously presented heterogeneous reaction scheme. The detection limits for arsenic, selenium, and antimony were 1.2, 3.7, and 10 μg/l, respectively. The calibration graphs were linear from the detection limit up to 125 μg/l for arsenic, 150 μg/l for selenium, and 250 μg/l for antimony. The relative standard deviations at concentration levels of 10 and 100 μg/l were 1.8 and 0.7% for arsenic and 2.3 and 1.2% for selenium. Corresponding figures for 50 and 100 μg/l antimony were 2.5 and 1.6%.