Tritium (³H) is a vital tracer in isotope hydrology and environmental studies, necessitating accurate and precise low-level detection methods. This study evaluated the performance of the new Hidex ULLA Liquid Scintillation Counter for ultra-low-level tritium assays in environmental water samples. The optimization of instrumental parameters enabled accurate and reproducible 3H measurements. Test samples (10–132 TU) were analyzed in triplicate, including 3H-free water and NIST traceable standards. The results demonstrated excellent accuracy and reproducibility, as evidenced by Zeta scores and linearity across these tritium concentrations. The system's stability over extended measurement time periods ensured reliable performance for routine ultra-low level 3H analysis. The optimization of key parameters such as coincidence time and region of interest (ROI) achieved the highest figure of merit (FoM) by using a 25 ns coincidence time and the DigPb open mode for optimal spectral fittings. Detection limits (DL) based on ISO11929 recommendations were significantly improved but highlighted the need for sample enrichment for most low-level environmental applications, with LSC counting times preferably exceeding 1000 min. Overall, the Hidex ULLA Liquid Scintillation Counter proved to be a robust and reliable instrument suitable for ultra-low-level tritium detection in environmental water samples.