A candidate reference measurement procedure (RMP) for serum theophylline via isotope dilution liquid chromatography-tandem mass spectrometry (LC-MS/MS) was developed. With a single-step precipitation pretreatment and a 6-min gradient elution, the method achieved baseline separation of theophylline and its analogs on a C18-packed column. A bracketing calibration method was used to ensure repeatable signal intensity and high measurement precision. The intra-assay and inter-assay imprecisions were 1.06%, 0.84%, 0.72% and 0.47%, 0.41%, 0.25% at concentrations of 4.22 µg/mL (23.40 µmol/L), 8.45 µg/mL (46.90 µmol/L), and 15.21 µg/mL (84.43 µmol/L), respectively. Recoveries ranged from 99.35 to 102.34%. The limit of detection (LoD) was 2 ng/mL, and the lowest limit of quantification (LLoQ) was 5 ng/mL. The linearity range extended from 0.47 to 60 µg/mL (2.61-333.04 µmol/L). No ion suppression and carry-over (< 0.68%) were observed. The relative bias for this candidate RMP that participated in 2023 External Quality Control for Reference Laboratories (RELA) conducted by the International Federation of Clinical Chemistry (IFCC) was within a range of 0.17 to 0.93%. Furthermore, two clinical immunoassay systems were compared with this candidate RMP, demonstrating good correlations. The results of the Trueness Verification Plan indicate significant differences among routine systems, highlighting the need for standardization efforts. The developed candidate RMP for serum theophylline serves as a precise reference baseline for standardizing clinical systems and assigning values to reference materials.