A scoping study of a commercially available resin selective for aqueous plutonium (Pu), AnaLig® Pu-02, modified with scintillator was investigated as a scheme to simultaneously concentrate and detect Pu in aquatic matrices. The extractive scintillating resin was comprised of a silica base, functionalized for plutonium extraction, grafted with plastic scintillator of polyvinyl toluene (PVT) and 2-(1-naphthyl)-4-vinyl-5- phenyloxazole (vNPO) fluor. Scintillator was incorporated onto the AnaLig® Pu-02 resin in a two-step process of silanization followed by surface-polymerization. Successful modification was facilitated by grinding the resin beads prior to silanization to expose cleaved silica surface sites appropriate for scintillator grafting. The modified resin was subjected to initial characterization of batch uptake and radioluminosity measurements where a total detection efficiency of 32.5% was observed. The modified resin was then subjected to pH 1 simulants containing environmental relevant groundwater constituents of varying concentration. Concentrations of 0.001M Fe(III) interfered with Pu uptake, while concentrations of up to 0.01M Ca(II) and 0.001 mM concentration of natural uranium and thorium had minimal influence on plutonium uptake. A translucent column packed with the modified AnaLig® Pu-02 was placed in a commercial flow-cell radiation detector for real-time detection of plutonium; a total detection efficiency of 20.4% was achieved for on-line measurements. The modification of AnaLig® Pu-02 results in a minimum detection limit capable of meeting the EPA limit for gross alpha activity in drinking water given a sufficient counting time of 15 min and approximately 300 mL of solution volume.
Read full abstract