An orthogonal acceleration time-of-flight mass spectrometer has been constructed with a matrix-assisted laser desorption/ionization source. The instrument has been set up in a preliminary configuration for an assessment of the viability of the orthogonal acceleration geometry. Measurements of resolving power and statistical evaluation of the mass accuracy/precision are reported here for ions in the range m/z 614 to 1351. The preliminary results indicate that resolving powers in the range of 2800 to 3600 (full width at half maximum) are readily achievable without the use of an ion mirror. A two-point external calibration of the mass scale was found to be stable over several days. The precision of the mass measurement has been estimated statistically in an error analysis involving about 40 measurements in the mass range specified above. The distribution of errors was found to be normal (linear normal probability plot) with a mass accuracy of about 0.1 to 70 ppm (average error) depending on how the calibration was performed, while the precision is found to be 90 ppm (standard deviation) irrespective of calibration method. The observed errors are consistent with calculations based on the resolving power and the 2.5 ns digitizer frequency used in this work. Measurements are restricted in mass range with the current instrument because it has a low ion energy (3 kV) accelerator and detector configuration that is linked to the accelerator by a long narrow drift-tube (1.5 m×20 mm). A new large-detector configuration with a 20 kV accelerator and a wider drift region is briefly described.