The 2 µm-wavelength thulium laser is an effective cutter during partial arytenoidectomy, but thermal trauma can damage adjacent laryngeal tissue. Pulsing laser energy may reduce trauma when compared to continuous-wave cutting. This study measured temperature changes, thermal trauma, and time to complete partial arytenoidectomy, with and without pulsing, in an ex-vivo calf model. Tissue temperature and time to complete a trans-cartilaginous cut were measured during partial arytenoidectomy on ex-vivo calf vocal folds (N = 24) using a thulium laser in continuous-wave (CW, N = 12) and pulsed-wave (PW, N = 12) modes. Energy was 5 W for CW and PW cuts; pulse-widths were 250, 500, and 750 ms. Thermal damage was analyzed histologically by measuring the depth of lactate dehydrogenase (LDH) inactivation perpendicular to the laser-cut edge at the vocal process. Paired t-tests compared CW and PW modes. Change in temperature was lower using CW (6.5°C) compared to PW modes (250 ms = 18°; 500 ms = 16°; 750 = 19°; P < .05). Trans-cartilaginous cuts were completed faster using CW (37 seconds) compared to PW (250 ms = 136 seconds; 500 ms = 61 seconds; 750 = 44 seconds; P < .05), and both modes delivered the same total Joules. The average depth of LDH depletion (thermal damage) was similar for all cuts. 1. Thulium laser cuts in continuous-mode unexpectedly produced less tissue heating yet created similar thermal damage than pulsed-mode cuts during simulated partial arytenoidectomy. 2. Trans-cartilaginous cuts were completed significantly faster in continuous-mode as compared to pulsed-mode cutting. 3. Pulsing the thulium laser does not minimize thermal damage compared to continuous wave cutting during thulium laser-assisted partial arytenoidectomy.