An ion micro-analyzer (Cameca IMS 300 with a TD 1000 computer unit) has been calibrated for quantitative point determination of fluorine in apatites, including tooth and bone material. Samples with macroscopically measured F concentrations were used as external standards. The samples were bombarded with O− primary ions. Secondary positive ion currents were registered as mass spectra containing mass numbers 19, 20, 31, 44, 59, 60 and 61. The latter mass numbers were used in isotopic correction to subtract oxide and hydroxide contributions from the fluoride 59 peak. For each bombarded spot, the ratios19F+/40Ca2+, respectively59CaF+/44Ca+, were plotted in a log-log graph against40Ca2+/44Ca+, the abscissa providing a measure of ionization efficiency. The plots yielded two sets of parallel straight lines, with ordinate values proportional to the respective F/Ca concentration ratios, in agreement with earlier theory. Each set of lines constitutes a calibration graph, where the fluorine concentration at an investigated spot is obtained when substituting the measured ion current ratios. With biological apatites, account must also be taken of effects due to, among other things, adsorbed humidity, other contaminations, and non-stoichiometry. When all precautions are taken, the accuracy of the determined F contents in the point analysis of an area some 100 μm in diameter can be expected to be better than 6% at 1500 ppm F, and better than 25% at 100 ppm F. The calibration is effective down to about 20 ppm F. Applications have been carried out on samples of human tooth enamel.