BackgroundAxillary vein puncture (AVP) is a valid alternative to Subclavan vein puncture for leads insertion in cardiac implantable electronic device implantation, that may reduce acute and delayed complications. Very few data are available about ICD recipients.A simplified AVP technique is described. MethodsAll the patients who consecutively underwent “de novo” ICD implantation, from March 2006 to December 2020 at the University of Verona, were considered.Leads insertion was routinely performed through an AVP, according to a simplified technique. Outcome and complications have been retrospectively analyzed. ResultsThe study population consisted of 1711 consecutive patients. Out of 1711 patients, 38 (2.2%) were excluded because they were implanted with Medtronic Sprint Fidelis lead.Out of 1673 ICD implantations, 963 (57.6%) were ICD plus cardiac resynchronization therapy, 434 (25.9%) were dual-chamber defibrillators, and 276 (16.5%) were single-chamber defibrillators, for a total of 3879 implanted leads. The AVP success rate was 99.4%. Acute complications occurred in 7/1673 (0.42%) patients. Lead failure (LF) occurred in 20/1673 (1.19%) patients. Comparing the group of patients with lead failure with the group without LF, the presence of three leads inside the vein was significantly associated with LF, and the multivariate analysis confirmed three leads in place as an independent predictor of LF. ConclusionAVP, according to our simplified technique, is safe, effective, has a high success rate, and a very low complication rate. The incidence of LF was exceptionally low.The advantages of AVP are maintained over time in a population of ICD recipients.