BackgroundThe aim of our study was to compare implant survival rates of different total hip arthroplasty (THA) bearings in the Nordic Arthroplasty Register Association. MethodsAll conventional primary THAs performed between 2005 and 2017 in patients aged more than 55 years who had primary osteoarthritis were studied. Metal-on-highly cross-linked polyethylene (MoXLP), ceramic-on-highly cross-linked polyethylene (CoXLP), ceramic-on-ceramic (CoC), and metal-on-metal (MoM) bearings were included. The outcome was a revision. Kaplan-Meier (KM) estimates were calculated at 5 and 10 years. The risk for revision was analyzed using a flexible parametric survival model adjusted for nation, age, sex, femoral head size, and femoral fixation. ResultsA total of 158,044 THAs were included. The 5-year KM estimates were 95.9% (95% confidence interval [CI] 95.8 to 96.1) in MoXLP, 95.8% (95% CI 95.6 to 96.1) in CoXLP, 96.7% (95% CI 96.4 to 97.0) in CoC, and 93.9% (95% CI 93.5 to 94.4) in MoM. The 10-years KM estimates were 94.2% (94.0 to 94.5) in MoXLP, 94.3% (93.9 to 94.8) in CoXLP, 95.4% (95.0 to 95.9) in CoC, and 85.5% (84.9 to 86.2) in MoM. Compared with MoXLP, the adjusted risk for revision was lower in CoC (hazard ratio [HR] 0.6, CI 0.5 to 0.6), similar in CoXLP (HR 1.0, CI 0.9 to 1.0), and higher in MoM (HR 1.3, CI 1.2 to 1.4). ConclusionsWe found that MoXLP, CoXLP, and CoC bearings evinced comparably high implant survival rates up to 10 years, and they can all be regarded as safe options in this patient group. The MoM bearings were associated with clearly lower survivorship. The CoC bearings had the highest implant survival and a lower adjusted risk for revision compared with highly cross-linked polyethylene bearings.