Before this study, no institution in China had undertaken the calculation and official establishment of reference values for the reference air kerma rate associated with high-dose rate 192Ir sources used in brachytherapy. This research, carried out at the National Institute of Metrology (NIM) in China, has successfully established an 192Ir reference radiation facility. Consequently, it has achieved the absolute measurement of the reference air kerma rate for high-dose rate brachytherapy using 192Ir sources. For this study, a medical afterloader machine was acquired. The radiation source employed was a high-dose rate 192Ir source for brachytherapy, produced by HTA Co., Ltd. A reference graphite cavity ionization chamber with a volume of 100 cm3 was designed to serve as the ionization chamber for absolute measurements. The determination of various correction factors and physical constants was achieved through a method that combines Monte Carlo simulations with experimental techniques. The study successfully accomplished the absolute measurement of the reference air kerma rate for the high-dose rate brachytherapy 192Ir radiation source. The expanded uncertainty of the measurement results was 0.72% (k = 2). The relative standard deviation for the RAKR of the same 192Ir radiation source was 0.073%. This study marks a significant advancement in the field of radiation therapy in China, particularly in the accurate dosimetry for brachytherapy using high-dose rate 192Ir sources. The study will contribute to the BIPM.RI(I)-K8 international comparison, organized by the Bureau International des Poids et Mesures (BIPM), thus facilitating the international recognition of the measurement values.
Read full abstract