The optimal treatment strategy for proximal humerus fractures (PHFs) is debatable owing to the relatively high failure rate of locking plates. Optimizing implants may enhance the fixation stability of PHFs and reduce the rate of mechanical failures. We developed a finite element (FE) model to simulate the treatment of PHFs with Proximal Humerus Internal Locking System (PHILOS) plates. The model evaluated the average bone strain around the screw tips under vertical loading (as an alternative to the risk of cyclic screw cutout failure verified through biomechanical testing) to minimize this strain and maximize predicted fixation stability. After determining the optimal screw configuration, further FE analysis and in vitro biomechanical testing were conducted on both standard and optimized PHILOS screw orientation to assess whether the optimized plates have biomechanical advantages over the standard plates. The FE-based optimized configuration exhibited significantly lower bone strain around the implant than the standard PHILOS screw orientation (− 17.24%, p < 0.001). In both FE analysis and in vitro biomechanical testing, the optimized PHILOS plates achieved significantly lower average bone strain around the screws (p < 0.05), more uniform stress distribution, and greater structural stiffness (p < 0.05) than the standard PHILOS screw orientation. Our results show that biomechanical performance of the PHILOS plates can be improved by altering the orientation of the locking screws. This approach may be useful for future patient-specific design optimization of implants for other fractures.