Herein, a simple and rapid HPLC method for the determination of paracetamol milk-containing biorelevant media is proposed. The separation of the analyte from the milk-containing biorelevant media was accomplished isocratically using a mobile phase containing 25 mM phosphate buffer (pH = 3.0) and methanol, 80:20, v/v at a flow rate of 1 mL min−1. Following a protein precipitation-based sample clean-up, a thorough investigation of the effect of the precipitation reagent (methanol, acetonitrile, 10% v/v trifluoroacetic acid solution) on the analyte recovery was performed. The matrix effect was assessed in each biorelevant medium by comparing the slopes of the calibration curves of aqueous and matrix-matched calibration curves. The method was comprehensively validated using the accuracy profiles. The β-expectation tolerance intervals did not exceed the acceptance criteria of ±15%, meaning that 95% of future results will be included in the defined bias limits. The relative bias ranged between −4.5 and +3.9% for all analytes, while the RSD values for repeatability and intermediate precision were less than 2.7% and 3.0%, respectively. The achieved limit of detection (LOD) was 0.02 μg mL−1 and the lower limits of quantitation (LLOQ) were established as 10 μg mL−1, which corresponded to 2% of the highest expected concentration of paracetamol. The proposed scheme was utilized for the determination of paracetamol in dissolution studies of its 3D-printed formulation in milk-containing biorelevant media.