Reuse of medical devices poses risks concerning technical issues and patient safety. In this study, we aimed to examine the structural changes in catheters that occur due to the reuse with the aid of electron microscopy. The effects of hydrogen peroxide (HP) and ethylene oxide (EO) sterilization on four percutaneous transluminal coronary angioplasty (PTCA) catheters and control PTCA catheters were examined by scanning electron microscope (SEM). Each catheter sample was divided into four parts during the SEM examination, and a total of 20 pieces were examined. Catheters were reprocessed through every regular sterilization step and used solely for the study, not in patients. Statistical evaluations of histological scoring made on images obtained from scanning electron microscopic images were made using the GraphPad Prism 8 program. Electron microscopical examination showed that HP sterilization caused more robust and deeper lines compared to EO. These distortions increased directly with the increase in the reprocessing cycle. In EO, no significant damage was detected within five cycles in contrast to HP; however, the harmful effects of EO were seen over five cycles. Unprocessed samples had no damage. Outer and inner deterioration was significantly higher in the EO>5 group and HP>5 group than in the control group. However, the bacterial contamination score in the EO>5 group was higher than the control group. Our findings showed that HP and EO sterilizations caused some deterioration in the inner and outer surfaces of PTCA catheter samples. We recommend reprocessing using EO, the least damaging method, when necessary, and paying attention not to exceed five cycles when necessary.