Maps of local plasma emissivity of TJ-II plasmas are determined using three-array cameras of silicon photodiodes (AXUV type from IRD). They have assigned the top and side ports of the same sector of the vacuum vessel. Each array consists of 20 unfiltered detectors. The signals from each of these detectors are the inputs to an iterative algorithm of tomographic reconstruction. Currently, these signals are acquired by a PXI standard system at approximately 50 kS/s, with 12 bits of resolution and are stored for off-line processing. A 0.5 s discharge generates 3 Mbytes of raw data. The algorithm's load exceeds the CPU capacity of the PXI system's controller in a continuous mode, making unfeasible to process the samples in parallel with their acquisition in a PXI standard system. A new architecture model has been developed, making possible to add one or several processing cards to a standard PXI system. With this model, it is possible to define how to distribute, in real-time, the data from all acquired signals in the system among the processing cards and the PXI controller. This way, by distributing the data processing among the system controller and two processing cards, the data processing can be done in parallel with the acquisition. Hence, this system configuration would be able to measure even in long pulse devices.