To study the response of the ArcCHECK® device as VMAT and IMRT verification system. Various tests analyzing the linearity, the repeatability and the angular dependence of the device response, its dependence with the pulse repetition rate and the leakage losses were performed. The long-term response in dose measurements and the uniformity of the detectors conforming the system were controlled using a statistical process control program. The Elekta Infinity™ 6 and 15MV photon beams were used. The device showed excellent repeatability and linearity. The differences between the responses obtained for any pair of angular incidences were less than 2%. The absorbed dose increased by 3% when the pulse repetition rate varied from 50 to 600MU/min. Results are in overall agreement with those found in previous works for the ArcCHECK®, in which a reduced number of the device diodes were analyzed, and for the MapCheck®, an older 2D device that used the same diodes. Charge losses were found to be negligible except for some of the diodes of the device. The statistical process control program is a very useful tool to control the correct functioning of the device in the long term. The results of the analysis carried out indicate that the working and stability conditions of the ArcCHECK® device are adequate for its purpose. The dependence with the pulse repetition rate should be considered when VMAT or similar treatments are evaluated. A control program for the statistical monitoring of the device would be desirable and useful.