One of the fundamental attributes of a liquid chromatography column is its stability when exposed to acidic and basic mobile phases. However, there have been relatively few reports to date on the stability of hydrophilic interaction chromatography (HILIC) columns. Here, we report the results of stability evaluations carried out for HILIC columns packed with ethylene-bridged hybrid or silica particles using accelerated conditions, employing a 100% aqueous pH 11.3 ammonium bicarbonate mobile phase at 70 °C. Under these conditions, the primary mode of column failure was a loss of efficiency due to the formation of voids resulting from the hydrolysis of the particles. We investigated the dependence of stability on the surface area of both unbonded and sulfobetaine-bonded ethylene-bridged hybrid stationary phases. The results show a clear trend of stability increasing as the surface area decreases. Several commercially available HILIC columns that are recommended for use with high-pH mobile phases were also evaluated. The results show times to 50% loss of the initial efficiency ranging from 0.3 to 9.9 h. Columns containing unbonded, sulfobetaine-bonded or diol-bonded ethylene-bridged hybrid stationary phases had longer lifetimes than amino-bonded silica or sulfobetaine-bonded, hybrid-coated, superficially porous silica columns.